Water pressure booster

PostPosted: Mon Nov 05, 2012 5:35 pm
by lwparchive
I’m looking to resolve an issue I have with the sterilizer in the microbiology lab. It is a prevacuum unit and requires 30-50psig, dynamic, for the vacuum system to operate effectively and uses 15gpm at peak consumption. The city water line to the unit only supplies 25-35psig currently. I am able to run cycles, however, a QC test I do weekly cannot go
to completion due to inadequate water pressure to the vaccum system. I was hoping to get more information about the water pressure booster system to see if it will provide me with the adequate pressure I need.

Thank you,

Lisa D

Re: Water pressure booster

PostPosted: Thu Dec 20, 2012 2:04 pm
by dlamp67
I would look at the Gould's Aqua Boost systems. You can find detailed specs on our website.